  3. LG E650 B31 - Invalid Tag
By thiesens , 2 March, 2021
Hi Gurux
I am trying to connect to a meter (L&G E650 B31), but i get "Invalid Tag".
Trace:
08:52:54
<HDLC len="9" >
<TargetAddress Value="1" />
<SourceAddress Value="10" />
<FrameType Value="93" />
<Snrm>
</Snrm>
</HDLC>
08:52:54
<HDLC len="1D" >
<TargetAddress Value="10" />
<SourceAddress Value="1" />
<FrameType Value="73" />
<PDU>
<Ua>
<MaxInfoTX Value="128" />
<MaxInfoRX Value="62" />
<WindowSizeTX Value="1" />
<WindowSizeRX Value="1" />
</Ua>
</PDU>
</HDLC>
08:52:54
<HDLC len="2A" >
<FrameType Value="10" />
<PDU>
<AssociationRequest>
<ApplicationContextName Value="SN" />
<InitiateRequest>
<ProposedConformance>
<ConformanceBit Name="Read" />
<ConformanceBit Name="Write" />
<ConformanceBit Name="UnconfirmedWrite" />
<ConformanceBit Name="InformationReport" />
<ConformanceBit Name="ParameterizedAccess" />
</ProposedConformance>
</InitiateRequest>
</AssociationRequest>
</PDU>
</HDLC>
08:52:54
<HDLC len="1D" >
<PDU>
<Ua>
</Ua>
</PDU>
</HDLC>
08:52:54
<HDLC len="36" >
<FrameType Value="30" />
<PDU>
<AssociationResponse>
<AssociationResult Value="00" />
<ResultSourceDiagnostic>
<ACSEServiceUser Value="00" />
</ResultSourceDiagnostic>
<InitiateResponse>
<NegotiatedConformance>
</NegotiatedConformance>
<NegotiatedMaxPduSize Value="0960" />
<VaaName Value="FA00" />
</InitiateResponse>
</AssociationResponse>
</PDU>
</HDLC>
Log:
GXDLMSDirector 8.2.2102.2602
Log created 08:50:27
08:50:41 Initializing Network connection.
08:50:43 Send SNRM request.
7E A0 07 03 21 93 0F 01 7E
08:50:43
7E A0 1E 21 03 73 C3 7A 81 80 12 05 01 80 06 01 3E 07 04 00 00 00 01 08 04 00 00 00 01 07 22 7E
08:50:43 Parsing UA reply succeeded.
08:50:43 Send AARQ request.
7E A0 2B 03 21 10 FB AF E6 E6 00 60 1D A1 09 06 07 60 85 74 05 08 01 02 BE 10 04 0E 01 00 00 00 06 5F 1F 04 00 1C 03 20 FF FF 80 AB 7E
08:50:43
08:50:43 Disconnect request
08:50:43
7E A0 37 21 03 30 6C 7C E6 E7 00 61 28 A1 09 06 07 60 85 74 05 08 01 02 A2 03 02 01 00 A3 05 A1 03 02 01 00 BE 0F 04 0D 08 00 06 5F 1F 04 00 18 02 20 09 60 FA 00 0A A8 7E
System.Exception: Invalid tag.
5 years 6 months ago
Your meter is sending UA reply twite.
http://www.gurux.fi/GuruxDLMSTranslator?translate=7EA01E210373C37A81801…
This looks strange. Can you read the meter with GXDLMSDirector? Have you modified the source code?
thiesens
Hi Mikko
Hi Mikko
I cannot read the meter. I get the error when i try to connect the meter.
The source is not modified.
There is a sub-meter via RS485, could that be a reason?
/Henrik
That other meter might be the reason. Try to connect using the serial number of the meter.
Change "Address Type" from Default to SerianNumber and set the serial number.
thiesens
Hi Again.
Hi Again.
That worked perfectly, thanks.
The trace is nice. can I view it somewhere for the Gurux.AMI?
You should see traces as a default.
You can check this. Find "TraceLevel" from appsettings.json.
It should be "Verbose". You can see send and received data in the command prompt.
thiesens
Thanks for your help.